We have a few methods in our lab that uses different diluents for standards and samples. Some use acetonitrile while others use dimethylacetamide.

We segregate our columns based on what diluent (acetonitrile used only on acetonitrile method). Does it really matter using the same column with different solvents. Column is a db5 by the way.
Sometimes you can get a solvent effect at the head of the column. Sample/solvent vaporizes and then condenses at the head of the column. If the stationary phase and the solvent are not compatible, you can get smearing of your solvent and analytes (tailing of the peaks). To me, that's the only concern. The 5-phase is rugged. The 2 solvents you mention should be fine for that phase. I presume you're running a temperature program in your oven. Make sure you go hot enough to get all of the solvent out before the next injection.
Bake out between different methods. Got ya thanks
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry