Hello:

I am running an 1100 with a RP Column and Isocratic flow.

Water: ACN :: 92:8 + TEA and pH Adjustment

I recently made a new very precise solvent mix. After degassing, I started up let it run for about 20 mins and then injected the sample. The retention time of the main peak and impurities was as expected. A single impurity peak was more or less flat but acceptable.

Now in the second injection, the main peak took almost 2wice as long to get eluted. The main peak, however was still very crisp (not spread out). In the second injection, I also got very crisp impurity peak.

Now I observe that there is a small reduction in the pressure, this could be due to improper degassing (?), however how could this lead to crisp impurity peaks and delayed main peak. This has happened to me earlier also. Back then we were testing the method out so, we didnt give much attention.

Since this is a USP method, the retention time of the main peak is approximately known.

I have another question: I have a binary pump so, some times I run the water in one pump and ACN in the other. This seems to have provided very stable pumping. However since this is an isocratic method. I mix the 2 some times and pump in one pump.

I always thought, it is better to mix earlier and pump. However in this case, it seems to be leading to retention time variations. Inspite of vigerous shaking and then degassing, is it possible that a pre-mix solvent is not homogeneous ? Nothing else could logically explain the sudden shift in retention time ?

Thanks for your thoughts.
1. Did you purge the unused channel of your binary pump?

2. Troubleshooting can involve measuring the flow rate, or putting your pre-mixed mobile phase in both reservoirs even though just pumping from one.
Since this is a USP method....
3. With a USP method, anything can happen. Yes, I do know these are "offiical methods"... let's just say that some are way better than others...
Hello

First of all I'd run leak test - it is great diagnostic tool. Your problem sounds like issue with valve/valves (AIV or outlet ball valve) in the pump.

Regards

Tomasz Kubowicz
3 posts Page 1 of 1

Who is online

In total there are 23 users online :: 1 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google Adsense [Bot] and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry