Turn off PDA 2996 detector

Posted: Wed May 20, 2015 1:22 pm
by jmsalcines@gmail.com
Hi everybody,

We run HPLC samples under Empower Pro 2 software, with an Alliance Waters 2695 Separation Module, with a Waters 2996 photodiode array detector.

Well, the issue is that we'd like to know how to program an instrument method to turn off the lamp automatically after finishing the quickset to save lamp life.

Any suggerence would be very much appreciated.
Thanks

Re: Turn off PDA 2996 detector

Posted: Wed May 20, 2015 10:23 pm
by Mritari
Make a shutdown method.

Rename an existing or make a new instrument method named "shutdown". Under the PDA's tab under events, @ time 0 event out number should read "event lamp" and event should read "off".

If you want to taper the flow or turn off the pump you can set that too.

When you run a sample set method and hit start, there should be a dropdown menu for shutdown method, set that to method you just created.

Re: Turn off PDA 2996 detector

Posted: Thu May 21, 2015 10:43 am
by jmsalcines@gmail.com
Thank you very much for your quick answer, Mritari.

That's just what we did: new shutdown method with event lamp off, trying time 0, 1, 5... but it doesn't work at all turning off the lamp. Although it works well setting the flow to 0,2 mL/min. None of our 3 PDA seems to obey the turning off the lamp issue.
Make a shutdown method.

Rename an existing or make a new instrument method named "shutdown". Under the PDA's tab under events, @ time 0 event out number should read "event lamp" and event should read "off".

If you want to taper the flow or turn off the pump you can set that too.

When you run a sample set method and hit start, there should be a dropdown menu for shutdown method, set that to method you just created.

Re: Turn off PDA 2996 detector

Posted: Thu May 21, 2015 10:53 am
by jmsalcines@gmail.com
oops, we got the problem fixed:

there are some connectors in PDA rear's part that connect through wires to the Alliance system and controls the event signals.

That's a hardware issue.

Re: Turn off PDA 2996 detector

Posted: Tue Aug 25, 2015 1:17 pm
by pharm1
hi jmsakines,

I working on the same issue. Could share which outputs from the alliance you connected to the which event in positions on the 2996.

I have managed to get empower shutdown to turn off my FLR detector, but not working for the 2996.

Thanks

Re: Turn off PDA 2996 detector

Posted: Thu Jun 23, 2016 5:07 pm
by carmonaperez
oops, we got the problem fixed:

there are some connectors in PDA rear's part that connect through wires to the Alliance system and controls the event signals.

That's a hardware issue.
What connections did you used? How you wired it?