head space vs transfer line
Posted: Tue May 05, 2015 10:18 am
by chem111
hello,
what is the difference between a regular CTC headspace vs transfer line?
And is it possible to adopt a transfer line method to a regular headspace.
Thank you,
Michael.
Re: head space vs transfer line
Posted: Tue May 05, 2015 10:32 am
by CE Instruments
One mimics a manual method with a gas tight syringe the other uses a loop and valve. Yes you can usually swap between the two methods without changing the preparation times, agitation etc.
Re: head space vs transfer line
Posted: Tue May 05, 2015 11:03 am
by chem111
Thank you very much
Re: head space vs transfer line
Posted: Tue May 05, 2015 11:19 am
by chem111
I have another question:
USP sample preparation for residual solvents specifies that the sample is weighted into a volumetric flask and then transferred to HS vial.
Will simply transferring the sample directly to the HS vial and adding water to achieve a similar concentration be any different?
Re: head space vs transfer line
Posted: Tue May 05, 2015 11:55 am
by Peter Apps
I have another question:
USP sample preparation for residual solvents specifies that the sample is weighted into a volumetric flask and then transferred to HS vial.
Will simply transferring the sample directly to the HS vial and adding water to achieve a similar concentration be any different?
How different it is will depend on how similar the concentrations are. In particular how do you plan to measure the volume of water, and/or the final volume of the solution ?
Peter
Re: head space vs transfer line
Posted: Tue May 05, 2015 12:07 pm
by kubowicz.tomasz
I have another question:
USP sample preparation for residual solvents specifies that the sample is weighted into a volumetric flask and then transferred to HS vial.
Will simply transferring the sample directly to the HS vial and adding water to achieve a similar concentration be any different?
How different it is will depend on how similar the concentrations are. In particular how do you plan to measure the volume of water, and/or the final volume of the solution ?
Hello
Headspace vial is not a volumetric flask
Regards
Tomasz Kubowicz
Re: head space vs transfer line
Posted: Sat May 23, 2015 1:24 pm
by wraymogg
Hi,
use class A volumetric flasks to measure the water.
I sterilize the HS vials at 600 degrees so even if they were graded the volume might change after sterilization.
Regards,
Vlad