I am using a 5890 series II GC with FID to analyze toluene and TCE with manual head-space injections. The column is an hp-5ms 30x.32x.25. My method is isothermal at 120 deg with column pressure of 60 kPa and flowrate of 30mL/min N2.
I am shooting each sample in triplicate and the area will be 1200 ( which is what it should be) then 2 minutes later it will be 9000 and the peak will be 30 seconds long. Sometimes, it will do it 3 injections in a row, or sometimes it will go 2 hours without doing it. Not sure what the issue is.
The flow-rate and and column pressure are set by manual dial. The flowrate is very touchy usually (brush the side and it will adjust), but has recently and sporadically become much less sensitive. It almost seems to stick
Thoughts on what to try to fix the problem?