Smeared or extended peak length issue GC-FID

Discussions about GC and other "gas phase" separation techniques.

6 posts Page 1 of 1
Hello,

I am using a 5890 series II GC with FID to analyze toluene and TCE with manual head-space injections. The column is an hp-5ms 30x.32x.25. My method is isothermal at 120 deg with column pressure of 60 kPa and flowrate of 30mL/min N2.

I am shooting each sample in triplicate and the area will be 1200 ( which is what it should be) then 2 minutes later it will be 9000 and the peak will be 30 seconds long. Sometimes, it will do it 3 injections in a row, or sometimes it will go 2 hours without doing it. Not sure what the issue is.

The flow-rate and and column pressure are set by manual dial. The flowrate is very touchy usually (brush the side and it will adjust), but has recently and sporadically become much less sensitive. It almost seems to stick

Thoughts on what to try to fix the problem?

Thank you,
are you saying the column flow is 30mL/min?
Yes, I believe so
The flow rate of 30 ml/min is the total flow through the inlet, of which about 1 ml/min is going to the column, giving a 30:1 split. It sounds as if the flow controller is faulty and that the wide peaks are from unintentional "splitless" injections. If you measure the flow out of the split, does it fluctuate ?

Peter
Peter Apps
Peter,

Yes the split does appear to fluctuate. I switched it to a splitless method since my sample is relatively low concentration. The column pressure will not stay constant for a any flow rate and the total flow rate still seems to drop or surge. Any thoughts on what to try next?
Agilents suffer from contamination of the split line and vent trap - you could try cleaning or replacing the liner and replacing the trap. Otherwise you have to suspect a faulty EPC, which would need to be replaced.

Peter
Peter Apps
6 posts Page 1 of 1

Who is online

In total there are 24 users online :: 0 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 24 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry