One of the alternatives to consider, if you would like to analyze basic compounds at moderate pH (2-7), s mixed mode chromatography. With Primesep columns you don't need to go to very high or very low pH. You will find a lot of buffers which you can use in this range (ammonium acetate, ammonium formate, formic acid, TFA, etc.). You can choose your buffer based on the detection technique or the objective of your experiments. Mobile phases with pH=2-7 are not going to do any damage to the pump, detectors, valves, etc.). You can retain basic analytes with dual mechanism-reverse phase and ion-exchange and obtain retention from void to indefinite retention by playing with nature if the buffer and pH of the mobile phase (within the range of 2-7)
In addition to ultra pure silica you have a "masking" effect of the stationary phase which is usually have lower pKa value then silica and is more controllable due to the universal nature of the ligand.
Here is the link for "Improving of Peak Shape of Charged Analytes"
http://www.sielc.com/pdf/SIELC_ImprovingPeakShape.pdf
and a link with a few reason why you don't need to go above pH-7 in order to retain basic analytes in controllable fashion.
http://www.sielc.com/pdf/SIELC_June_2005.pdf
Leader column manufacturers are making columns with very pure silica and doing end capping in order to improve peak shape (Zorbax and Luna in my experience showed very good peak shape).
IN a lot of cases regular LC column with ion-pairing reagent will give you good peak shape (tailing factor of 1.5 or less).
Regards,
Vlad