I am troubleshooting a Waters 715 Autosampler. The pressure transducer does not seem to be working correctly. I am not sure if it needs to be replaced or if an adjustment can be in the diagnostic service
menus. Does anyone know how to determine the diagnostic service code?

Thanks,

Mark
Mark Mitch
New England College
We do repair works for all models of Waters HPLC system including the 712, 715, 717 A/S. http://www.welltechinc.com
Thanks. Good to know someone still provides for these older autosamplers. I replaced the pressure transducer and the unit is working fine.