Detector sulfur (SCD355)
Discussions about GC and other "gas phase" separation techniques.
10 posts
Page 1 of 1
Have a gas chromatograph coupled to the detector nitrogen and sulfur (SCD355) and when using Agilent oven to 360 ° C, the ceramic tube of the SCD (sulfur detector) loses sensitivity. I wonder what might be causing this problem.
-
- Posts: 75
- Joined: Thu Aug 15, 2013 7:44 pm
With the little information supplied, it's likely a combination of 1) your column bleed is coking the ceramic tubes or 2) your sample is very "heavy" and is coking the ceramic tubes.
Has your column been properly conditioned so the baseline rise on the SCD at 360 C is minimal? The SCD is very sensitive to column bleed.
Has your column been properly conditioned so the baseline rise on the SCD at 360 C is minimal? The SCD is very sensitive to column bleed.
-
- Posts: 5
- Joined: Mon Mar 09, 2015 6:06 pm
Yes, the column was conditioned with 360 ° C. Someone already had the sensitivity loss problem in sulfur detector 355? The silica may be interfering in the detector sensitivity?
-
- Posts: 75
- Joined: Thu Aug 15, 2013 7:44 pm
What is your definition of conditioning the column? What are the specs of your column?
Who hasn't had a sensitivity loss problem with a 355 SCD at one point?
Who hasn't had a sensitivity loss problem with a 355 SCD at one point?
-
- Posts: 5
- Joined: Mon Mar 09, 2015 6:06 pm
The column is that I use Agilent - DB-5MS (Length - 30m / Diam - 0.250 mm / Film 0,10μm).
-
- Posts: 75
- Joined: Thu Aug 15, 2013 7:44 pm
How long was the column conditioned at 360C? And was the column connected to the SCD during conditioning?
I would suggest the following:
1) condition the column more, but have it disconnected from the SCD
2) replace the ceramic tubes in the SCD burner
I would suggest the following:
1) condition the column more, but have it disconnected from the SCD
2) replace the ceramic tubes in the SCD burner
-
- Posts: 5
- Joined: Mon Mar 09, 2015 6:06 pm
During the conditioning the column was not connected to the detector and SCD ceramic tubes are new. The problem is that I made standard injections repeatedly and the sensitivity drops and the value of the area decreases.
-
- Posts: 45
- Joined: Tue Oct 04, 2011 9:14 am
You could try turning the hydrogen off to the SCD between injections, while the GC is cooling.
-
- Posts: 5
- Joined: Mon Mar 09, 2015 6:06 pm
Every time the GC is not running hydrogen is off. I wonder if the silica may impair the performance of SCD and therefore lose sensitivity. Know someone who has been through this sensitivity loss problemo at high temperatures?
-
- Posts: 75
- Joined: Thu Aug 15, 2013 7:44 pm
Please read my first response again. Column bleed cokes the ceramic tubes causing impaired sensitivity. You must not have conditioned your column properly at your final run temperature. It is the column phase, not silica, that is the problem.Every time the GC is not running hydrogen is off. I wonder if the silica may impair the performance of SCD and therefore lose sensitivity. Know someone who has been through this sensitivity loss problemo at high temperatures?
What is the purpose of your method going to 360 C?
10 posts
Page 1 of 1
Who is online
In total there are 65 users online :: 1 registered, 0 hidden and 64 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: BB65 and 64 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: BB65 and 64 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science