 - Posts: 4
 - Joined: Tue Feb 24, 2015 12:15 am
 
1. Plug MS at the interface : Air water check passed (No leak in MS side)
2. Connect the transfer line column to the MS but plug the other end going into the splitter (Dean's switch) : Air water check passed (No leak in transfer line and interface nut)
3. Connect transfer line to the Dean's switch but plug all other outlets of the switch (viz GC column and FID). : Air water check failed. We start seeing N2 peak > 15-20 %
(and it varies with different runs) This made us conclude that there is a leak somewhere in Dean switch. But we could not identify the exact location by using Dust removal spray. So I have 3 options :
1. Go for on site agilent maintenance. (Very expensive and will take time)
2. Buy a new dean's switch. (Expensive and risky too as we are not 100% sure that it will fix the problem)
3. Continue with the small leak (around 15%) (Cheap but I dont know how will it affect our result and other parts of instrument)
Please, need advise.