When analyzing samples on an Agilent GC using Chemstation software some injections will show no signal except a very small (~300 on the y axis). The next injection will have a normal chromatogram and if we re-inject the same vial it will also have a normal chromatogram. The syringe, liner, septum and column have been changed and the problem still occurs. The more samples the instrument injects the more often it happens (the frequency will increase).
-
- Posts: 526
- Joined: Fri Feb 10, 2012 7:39 pm
So this issue only manifests with the autosampler, and not if you run such sequence using manual injection?
Then sounds like an autosampler issue, and if it appears to get better sometimes after "no peaks", seems to indicate a plugged syringe needle.
Are you using a needle wash either pre-or post injection? Repeated pumping and/or rinse steps?
What's the liquid level in the autosampler vial? Should be like one half to two-thirds the volume of the vial, fully filled can cause a vacuum effect.
Then sounds like an autosampler issue, and if it appears to get better sometimes after "no peaks", seems to indicate a plugged syringe needle.
Are you using a needle wash either pre-or post injection? Repeated pumping and/or rinse steps?
What's the liquid level in the autosampler vial? Should be like one half to two-thirds the volume of the vial, fully filled can cause a vacuum effect.
-
- Posts: 4
- Joined: Wed Jan 28, 2015 10:06 pm
We have changed the syringe and still having the same problem, the vials are 2/3 full and a methanol rinse is done pre and post (2 before and 2 after each sample injection).
-
- Posts: 880
- Joined: Fri Jan 04, 2008 3:03 pm
Make sure the rinse solvent works with your sample solvent. I had issues with methanol when iusing BSTFA as a sample solvent.
-
- Posts: 1113
- Joined: Wed Dec 22, 2010 10:17 pm
Never use a protic solvent as a rinse with a silyating reagent. Use acetonitrile, acetone, MeCl2/Chloroform, pyridine (viscous need a viscocity delay), MEK, ethyl acetate, or a hydrocarbon (hexane, hepatane, pentane etc). Silylating reagents and the analyte products are hydrolyzed by water or any protic solvent (alcohols or amines)
-
- Posts: 1113
- Joined: Wed Dec 22, 2010 10:17 pm
That reminds me are your samples very viscous. You might need to enable a viscocity delay or the injection could be very noreproducible due to insufficient time to draw in sample or getting air in. Also make sure there are enough sample strokes (I usually have 5 or 6).
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
Hi Blakey
It would be really, really helpful if you told us what it is you are analysing, and the conditions under which you are running - flow rates, temperatures, injection volume, inlet liner type, column, sample solvent etc etc etc.
Otherwise we have to troubleshoot by guesswork, which is slow and unproductive.
Peter
It would be really, really helpful if you told us what it is you are analysing, and the conditions under which you are running - flow rates, temperatures, injection volume, inlet liner type, column, sample solvent etc etc etc.
Otherwise we have to troubleshoot by guesswork, which is slow and unproductive.
Peter
Peter Apps
7 posts
Page 1 of 1
Who is online
In total there are 178 users online :: 1 registered, 0 hidden and 177 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Bing [Bot] and 177 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Bing [Bot] and 177 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science