by
Jake » Thu Feb 12, 2015 8:33 pm
We're having an intermitent error, where after completing all the syringe wash cycles, the syringe/arm just critically stalls over the waste. The message on the PAL display is "head communication error." The log in OpenLab EzChrom under PAL control just says "error." This used to happen rarely, but the error is happening more frequently. This is on an Agilent 7890A with an Agilent-branded PAL. We have a service contract with Agilent, but they're really dragging their feet on this and we need it fixed. The instrument handles nearly 200 samples per day, so we're falling hopelessly behind. One engineer was on-site, checked everything and didn't find anything wrong. Problem persists. An Agilent PAL expert later spent the afternoon here and basically all he did was update the firmware from 4.1.4 to 4.3.0. After spending a few hours re-building all our custom objects and re-training (thanks Agilent!), the same error occurred a few hours into that night's run.
Has anyone had and solved this error? Any suggestions? We need to get this fixed, and the next engineer isn't scheduled to be here for another two weeks. The instrument has 108,000 injections on it. It's configured with the larger wash and waste containers. The stall almost always happens after completing all the syringe wash cycles, although it did stall once in the middle of a wash cycle. But the error always occurs during syringe washing. We have another Agilent, a Agilent 7890B configured similarly, with about 51,000 injections on it, and it just generated this error for the first time few nights ago.
Thanks for any help.