A Follow-up to Gerhard Kratz:
Why do you want to switch to another column? Hypersil Gold is available in all imagineable dimensions and particle sizes. So you just have to scale down the column using the abovementioned tools. Try to keep flow rate > 0.5 mL/min if you use a quarternary pump. Binary pumps should be OK down to 0.25 mL/min (my experience, YMMV).
Kind regards,
Jörg
The reason we don't just switch to another size of Hypersil Gold is because of money. The lab I work in is at a small university with a large undergraduate program, so money can be an issue at times. The reason for the switch to the Poroshell 120 specifically is because it was given to us as a sample by a very generous sales rep.
What kind of resolution are you getting between your "critical pair(s)"? Have you tried optimizing your existing system at all for throughput?
Tell us more about your methods
What are "A" "B"?
Your gradient program?
Flow rate?
Column particle diameter?
When it comes to resolution, I have nice sharp peaks (Hypersil GOLD) depending on how pure the synthesis part of our lab can get the sample (for metal complexes).
For methods, this is our current set up.
A = 0.1% formic acid in H
2O
B = 0.1% formic acid in MeOH
GRADIENT
The gradient for metal complexes is:
Time(min)
Hypersil GOLD-----A(%)--------B(%)
---0.00----------98----------2
---2.00----------95----------5
---5.00----------70----------30
---15.00---------70----------30
---20.00---------40----------60
---30.00---------5-----------95
---35.00---------98----------2
---40.00---------98----------2
The same gradient adjusted for the PoroShell is as follows (taken from the advice given by tom jupille - Thank you ).
Time (min)
PoroShell 120--------A(%)--------B(%)
---0.00-------------98---------2
---0.80-------------95---------5
---2.00-------------70---------30
---6.00-------------70---------30
---8.00-------------40---------60
---12.00------------5----------95
---14.00------------98---------2
---16.00------------98---------2
I have since tried this method on the PoroShell 120 but found my standards coming off too late. The standard I made is a 100µM solution (in DMSO) of some ligands commonly used in our lab, incase anyone is wondering. Since the PoroShell 120 gradient above let the standards come off too late in the run, I changed the gradient to that seen below
Time (min)
PoroShell 120--------A(%)--------B(%)
---0.00-------------98---------2
---0.80-------------90---------10
---2.00-------------70---------30
---6.00-------------40---------60
---8.00-------------20---------80
---12.00------------5----------95
---14.00------------98---------2
---16.00------------98---------2
I ran a photoejecting experiment (which is what I do with the HPLC and our metal complexes) using the method immediately above. The placement of the peaks before, during, and after ejection were relatively the same as using the Hypersil GOLD column. I now have a problem with peak tails (I believe that is the proper term, when the peak trails at the end). I only just got the result yesterday and I currently trying to correct for it but tips are welcome.
FLOW RATE
The flow rate for the Hypersil GOLD has always been set at 1.00 mL/min (pressure stayed at ~120 bar while on standby, and ~170 bar when running a sample) , and I have been setting the PoroShell 120 at 0.500 mL/min (pressure is ~200 bar on standby and hovers ~300 bar while running a sample). The subject of pressure begs a question I will make in a different post.
PARTICLE SIZE
The particle size for the columns used are as follows:
Hypersil GOLD = 5 µm
PoroShell 120 = 2.7 µm
I apologize for any mistakes or explanations that may not be clear. Let me know and I can try to elaborate and/or explain differently.
Thank you.