It has been several years since I have worked with the chemstation software for an Agilent HPLC. When the detector (Fluorescence) turns on it goes from 0 up to 17 on the y axis and that is where it stays during the analysis. I would like to set it to read at 0. I know it does not change the chromatography any, but I would still like it at 0. I've looked on line and did not see anything.
It looks like you can add a baseline zero event in the timetable under the FLD method parameters.

Depending on the specifics of your method, a baseline that high might be caused by light scattering from a dirty flow cell or optics, but as you said, that should not affect your chromatography if the baseline is stable.
David Hobbs
Instrument Services Specialist
Pace Analytical Services, Inc.
david dot hobbs at pacelabs dot com
2 posts Page 1 of 1

Who is online

In total there are 408 users online :: 0 registered, 0 hidden and 408 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 408 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry