We have a CP-3800 and for some reason it has lost configuration with the modules. The icon for the GC is lost and there are no available modules in the task bar. How do I get them back?
-
- Posts: 14
- Joined: Thu Jun 26, 2014 9:35 am
Cindy,
Are you using GC MS? Which is your software?
If you are using MS WS there is an option in the "instrument " to setup Ethernet Communications and you can add your IP address of gc in respective column and you will get connected.
I'm sure you are out of this problem by this time .
Jihad
Kuwait
Are you using GC MS? Which is your software?
If you are using MS WS there is an option in the "instrument " to setup Ethernet Communications and you can add your IP address of gc in respective column and you will get connected.
I'm sure you are out of this problem by this time .
Jihad
Kuwait
2 posts
Page 1 of 1
Who is online
In total there are 222 users online :: 0 registered, 0 hidden and 222 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 222 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 222 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science