I was surprised to see the statement that says: "For gradient separations, changes in length, column inner diameter and particle size are not allowed".
This means we can only change isocratic methods and yet I know from experience that gradient methods can be successfully and easily converted when the gradient steepness based on column volumes is kept the same for each gradient step and making an allowance for difference in delay volumes between systems.
Has anyone else got views or comments on this?