Trace GC Ultra column K factor low

Discussions about GC and other "gas phase" separation techniques.

6 posts Page 1 of 1
Hi All,

I have a question about column evaluations on a Trace GC Ultra from Thermo. Our column (30m, 0.25mm id, Restek #13623) passes the leak check but its K factor is much lower than expected (0.8 instead of the expected 1.7). I also understand that a meaningful leak check might depend on a column evaluation performed with a truly leak-tight system.

I checked all connections with an electronic leak detector and found no leaks.

Does anyone know what the underlying problem could be? Is this a problem at all?

Thank you so much!

Arne
It's about right for a 0.32mm id column :?
Low K factor indicates a lower column back pressure.
How new is the septa ?

If the GC thinks youir column is wider than it is then it will set too low a pressure when calculating the flow so constant flow modes will give slower than expected chromatography. Have you injected a solvent to see if the time is later than expected ?
1. do you have vacuum compensation function turned off? this would decrease the K-factor as well
2. are you sure the column is not broken somewhere in the middle? any bubbles going out at the end of the column merged in a solvent?
Hi,

thank you so much for the replies!

CE Instruments:

I already noticed that it would be about right for a 0.32mm column.
I was also aware that a lower K factor indicates a lower backpressure.
I don't know what expected retention time should be since I recently took over the instrument after the previous operators left my place of work. RTs seem reasonably in agreement with what I see in published work. I will inject a bit of diethyl ether to determine the dead time and calculate whether that's appropriate - great idea.

trozen:

1. I have the vacuum compensation tuned on when the column is connected to vacuum and turned off when it isn't conencted to the mass selective detector. The obtained K factors are very similar.

2. I am sure that the column is not broken somewhere in the middle. I see bubbles come out in solvent at a rate that responds to changes in flow rate settings.



I am not sure what to do next...

Best,

Arne
One of the silver brazes on the injector had developed a leak. What an unlikely source of a leak! We had it re-brazed and the system is fine. Thanks for the input.
One of the silver brazes on the injector had developed a leak. What an unlikely source of a leak! We had it re-brazed and the system is fine. Thanks for the input.
Thanks for the feedback. One of the supposed advantages of back-pressure regulation of column head pressure is that the pressure is robust to leaks. It does not look as if that was the case here.

Peter
Peter Apps
6 posts Page 1 of 1

Who is online

In total there are 176 users online :: 0 registered, 0 hidden and 176 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 176 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry