LCQ Advantage ; No signal and failure on power supply

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
Please help, Our LCQ has problem and don't have any signal. i run diagnostic tests and has fail in power supply. (+28 is out of range it is reading 5.6v but -28V is pass).
I think this fail, effect on RF1 and appear multiple failure on;
Multipole Det RF;Fail
Multipole RF Mod: Fail
Multiple RF Amp: Fail.
so how can i solve problem on power supply +28V?
Let me first admit that I am no wizard on the older LCQ series instruments but I seem to recall that most of the voltages are somehow generated or routed through the System Control PCB.. The big board on the right side.. As the first thing I would check fuses on that board..

If you recently have had the system vented to clean the optics also check that you haven't caused a short related to the multipoles or maybe have missed a wire.. that could show up like this..

Good luck and let us know if you find the solution.
Kind regards
Leadazide
2 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry