The 5975B is not the best for gas analysis, and you have barely given us any info regarding injection etc. The traces all seem to show peaks at 3.5 minutes do you get separation ? Why can you ignore Argon ? From the scales should you not have a huge peak for mass 40? Could the Argon be supressing the response for Oxygen ?
Thank you very much for your answer.
Sorry if I didn't give enough detail, the sample is taken on-line, we have a probe connected to a transfer line and a vacuum pump. With the "gas sampling valve" (6-way rotary valve in the GC) we fill the 1 ml loop, and then injected in the injector with the Helium flow. The injector is working in split mode at 40:1 ratio. (I have tried splitless and different ratios and still the same problem).
What I sent before was the z/m 32 and 16 only but not the complete chromatogram. If I analyse the z/m 40 as you say I found a huge peak at about 3.5 min (is eluting at the same time as O2). I cannot separate with the instrument I have now (I would need a cryogenic stuff) the O2 from the Ar, so there is only 1 peak for this two. After that I use the convolution to "separate" it with the MS signal.
I can ignore Argon because I used as a diluent, I am not interested in the Ar behaviour, composition... It is an inert gas, and it should be constant during my process.
I don't think the Ar is suppressing the O2 response because when I analysed O2+Ar only, I cannot see any problem, and I have also diluted O2 with much more Ar and I can still detect it. It is only when I mix it with CH4 when I have the problem that it disappear.
Thanks for the advice that the 5975B is not the best for gas analysis, but it is what I found when I started this job, and right now we don't have a budget for this, so I will try to live with this, but I'll keep in mind your advice in case we can change it.
Once more thank you very much for advice, as I said I am totally lost with this.