I am wondering if anyone has any experience or seen anything like this in their vast years of experience
We have a new residual solvent HS-GC method for the determination of know solvents in an peptide API.
An unknown peak was obtained in comparative testing analysis and this was eventually identified as diethylamine (used in the process) at approximately 150ppm. This has been confirmed with GC-MS and 2 separate HS-GC methods.
The levels of diethylamine have also been determined using IC and the level is much lower at approximately 10ppm.
In trying to determine the true results we have seen that heating the sample for longer periods of time in the headspace oven has led to a large increase in the levels of diethylamine observed in our sample. We have passed this onto our chemists but they are unable to come up with a mechanism that would lead to the formation of diethylamine in the molecule and the impurities.
I am wondering if anyone may have seen anything like this previously.
Is there any experiments that we could try and perform to determine the true result?
Also would I be correct in assuming that for headspace analysis 30minutes equilibration should be more than sufficient for equilibration or could there be other mechanisms in place that would mean we need to give the sample much more time to reach equilibration?
The 3 main methods used are summarized below
IC: Ran on CS-17 column with Methane Sulfonica Acid and sample preppred in 5mMol Naoh
GC-MS: Ran on 624. Headspace oven at 120C for 30mins. Sample prepped in dimethylimidazolidione + 100uL NaOH
GC-FID: Ran on 624. Headspace oven at 80C for 60mins. Sample prepped in DMSO/NaOH.
Any help/input would be greatly appreciated.
Thanks in advance.