Could be, but when the solvent has been evaporated he will still have a layer of syrup in the bottom of the vial.He must mean an aqueous or other solvent (DMSO?) sugar solution. The whole idea of FET is that there is minimal matrix effects so the type and ammount of sugars from sample to sample should have no significant effect on results.
Impressive - at the LOQ you only have 5 ng of your analyte in the vial, so about 0.5 ng on the column if you are taking 1 ml headspace and transferring splitless.possibly depends on how much he diluted the sample. Since with FET you can only put a small amount of sample into the vial at most there will be some crystals at the bottom. I ended up going with FET for my residual ethanol method and am working on a diacetyl method. Most of my samples are also carbohydrate rich (they are starch encapsulated). I take 1 g sample to 10ml water with acetonitrile as my ITSD. I then put 5ul into a 9ml vial on my Tekmar 7000 and heat it to 120 deg C. LOQ is 10ppm in the vial using GC/MS SIM so with samples 100ppm.