Shimadzu SIL-20ac Trouble code: ERR TEMP SENSOR

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
The lab has an SIL-20ac... great machines. We're getting an ERR TEMP SENSOR code. Ignores all input. Still there after power cycling.
No mention of the temp sensor in our manual or any mention of this error code.

Any advice would be appreciated.

Thank you.
We checked the connection on the back of the SIL. It appeared to be exceptionally dirty. We cleaned it and touch wood all is functioning again.
Whew.