Issues:
-loss of sensitivity. Our LLOQ of 0.3 ng/ml is no longer visible in samples, nor in pure standards. I'll note that we perform regular maintenance on the inlet and source.
-terrible accuracy (precision isn't great but not bad). Our validation attempts have strangely had really great linearity of the cal curve (ascending order, first 10 vials), but the randomized QCs have wildly different RRs throughout the batch. We've run standards at the same concentrations and in the same order and, bizarrely, had the same issue of fantastic cal curve linearity but terrible accuracy of QCs. It's safe to rule out pipetting error for this issue.
-split peaks. Our analyte has split into 2, even 4 peaks sometimes. The internal standard occasionally has the same problem, but these issues only happen for a couple samples of each batch, including the recent "validation" of standards. The analyte abundance didn't decrease in that run. If anything it increased slightly. Compared to last month's performance, however, the abundance of internal standard has vastly decreased.
Things we've tried:
-we perform regular maintenance on the inlet, including putting in a fresh liner, trimming column, replacing gold seal, new ferrule.
-the column has been replaced but is the same kind we've used for the past year without these problems.
-the source has been cleaned. Many times.
-reagent gas: we've tried switching from methane to ammonia, but same problems.
-we're presently cleaning the split line, which is kind of gummed up, though we're operating in splitless.
If anyone has a suggestion I'd love to hear it. It must be a machine problem since standards produce the same issues, but we just can't figure out what. Thank you!