- Posts: 2
- Joined: Tue Dec 06, 2011 6:59 am
I have a strange problem going on which I would like to have you're opinion on.
The past two weeks I have been doing an Itraconazole analysis and from the beginning I'm having problems with the reproducibility. When doing 10 injections from the same reference solution I'm for example getting 4 injections with a peak area of 4600 mAU and 6 injections of 4400 mAU. The RT's are pretty stable.
The analysis is based on a monograph and has been validated internally. The following conditions are used:
System: Agilent 1100 with DAD
Column1: Polaris 10cm, 4.6mm, 3 um with guard column
Column2: Symmetry 10cm, 4.6mm, 3.5 um with guard column
Mobile phase A: Tetrabutylammoniumhydrogensulphate buffer (54 grams per 2 liter) (ion-paired)
Mobile phase B: CH3CN
Gradient elution
Injection volume: 5 ul
Detection: 225nm +/- 16nm
Because at first I thought it was an injector problem I replaced the following items in different stages: rotor seal, stator face, needle, needle seat and this didn't solve the problem. As a final resort I changed to another Agilent 1100 system.
Because at some point I got split peaks with the Polaris column I had to change it to a new one but this kind of column wasn't available and so my supervisor wanted me to change to the Symmetry column. I'm getting the same problem on that column. These columns are reserved for this analysis because of the ion-paired mobile phase.
A week before I got these problems I ran the same analysis just fine. Only thing I can think of now is that we are using a different jar of the buffer salt.
Anybody got any idea's?