In my last HPLC run with a TSK Gel CM-STAT cation exchange column which ran Thursday afternoon and overnight until Friday I have run in to instrument problems.
My first injection is just mobile phase A (10mM sodium phosphate) + 2uL carboxy peptidase.
The first chromatogram looked harmless like previous ones obtained last week and this week (where I am trying to deal with a retention issue, but that is another problem, and not covered in this question).
It is a 55 minute gradient with mobile phase B (10mM sodium phosphate + 100mM sodium chloride)
The method is described in more detail on page 6 here:
https://mc.usp.org/sites/default/files/ ... %200.2.pdf
"Procedure 3: LIMIT OF CHARGE VARIANTS (AFTER CARBOXYPEPTIDASE TREATMENT): CATION EXCHANGE
CHROMATOGRAPHY"
Then in the second chromatogram after approx. 16 minutes (early on the gradient) something goes ballistic…
In a split second it goes from around 0 mAU to 300 mAU and criss-crosses between 250 and 350 for the rest of the analysis.
Is it possible to post a picture of the chromatogram here? Wonder if this will work?
The rest of the sequence is similar intense baseline noise in all analyses.
I have tried the obvious during Friday, took out the column after it was washed through.
Put in a bit of peak tubing and flushed the system with MeOH20:80% first and then 100% MeOH. up to 5mL/min.
No change in baseline. I tried to pull air from the system lines and especially the A line seemed somewhat prone to air.
B also. D line (100% MeOH) and C line (20:80 MeOH:H2O) were both okay. Minimum air.
The lamp does not show any significant change in intensity. I have turned it off and back on again. No change.
I have tried to put back pressure on the flow cell to release possible air bubbles and the 100% MeOH should have helped as well. But there is no change.
The pressure is unchanged. No indication of leaks or blockage. No salt deposits on the syringe or the injection seal.
Any suggestions to explain how such a quick and massive deterioration of the output from the system is possible are very welcome...
Because it happens so suddenly, I was convinced it was “just” an air bubble trapped in the flow cell.
Now I am not so sure any more. Am I missing something? Could it be large amounts of air in the pumping system?
Could the degasser have blown a fuse?
My system is an old but normally quite reliable system.
Help...