If that would be right we would have 99,99% less problems in RP Chromatography! Different base silica, different bonding compounds etc. and have a look on the hydrophobicity tables published in books, comparing RP columns.
From one point of view, you're right, Gerhard. I do know that not all C18 phases were not born equal! However I do not think that changing from one C18 to another will ever bring to a total reversal or total randomization of the elution order. From what I've read and seen, the most orthgonal phases are C18 and pentafluorophenyl. Any good citation of literature is welcome.
This depends on what is sought. An effective 2D-HPLC for proteins/peptides (as in the Anal Chem paper) may or may not be useful for small molecules (as in the other cited papers).
Also this is true. The paper shows orthogonality of columns with a peptide mix. This does not imply that with another kind of analytes (whether it can be antibiotics, metabolome or any kind of small molecule) the orthogonality will be the same.
n the HILIC-HILIC 2D HPLC references you cited, polar solutes were sought. My guess is that HILIC-HILIC 2D HPLC would be an effective alternative for non-ionizable solutes that are too polar for RP-HPLC.
Yes, that's the whole point of the post! For non polar analytes, you can do RP-RP, for medium polarity you can do RP-HILIC, but nobody has explored yet HILIC-HILIC. What I was asking is:
- References from literature, if someone can find others
- Discussion about how one could achieve high orthogonality in HILIC.
Regarding the last point it seems obvious that one won't go very far using two bare silica columns. However there are many phases beyond bare silica, and there seems to be more difference in selectivity between them that between any tow RP phase (see
http://www.sciencedirect.com/science/ar ... 7311008491 and
http://www.sciencedirect.com/science/ar ... 7312012988). Ideally one would couple two colums that are opposite in the graph, obtaining high orthogonality, while still having the advantege of solvent compatibility (no peak distortion at injection, no sample dilution)