Here is what I've come up for for my residual ethanol method. I get two types of sample liquid flavors in soybean oil, propylene glycol, triacetin, or Neobee and spray dry encapsulated or pan dry encapsulated. I need to ensure they have minimal ethanol for halaal and my goal is to get down to below 100ppm in the samples for LOQ. I have a Tekmar 7000 deactivated sample pathway (unfortunately except for the cut end of the inlet carrier line) and a 6890N-5975MSD. Liner 1mmID. column Innowax or 1701 30m .25mm .25um film.
Here is what I am thinking. I've read Chromatographer1's advice regarding using a low sample volume in the vial. For diluent I am thinking of Using alpha-pinene. I have a ton available it has good oil and alcohol solubility and some/minimal water solubility with a Bp of 155 deg C. I might need a separate method with water for the encapsulated samples though.
Parameters
-100mg sample weighed into a headspace vial
-300ul of alpha pinene containing standard for standard addition
-9mL vials
Vial pressure 15PSI N2 supplied by 6890 aux EPC
Platen temp 80 deg
Loop temp 165 deg
Tline temp 175 deg
Sample equil 2 min
mix time 8 min
mix power 4 (max recommended for 9ml vial)
post mix equil 1 min
Pres time 0.2min
Pres equil .05min
Loop fil (1 ml) 0.25min
Loop equil 0.05min
inject 1 min
Inlet 200 deg C
3:1 pulsed split 3 ml column 9 ml split 2-3 septa purge so a total of ~15 ml/min transferline for 1 minute
Column flow 1ml/min after 1 minute
40 deg for 5 minutes +3 to 75 deg C +20 to 200 5min.
GC/MS SIM 45&46 m/z
I may have to loose the pulse if ethanol doesn't focus well.
Any comments?