Effect of heat exchange capillary length on retention time

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
What is the effect of length of heat exchange capillary on retention time
Depends a lot on column dimensions and flow rate.
If you go from 3 uL to 12 uL you will have little to no effect on a 250x4.6 mm column with flow rate 2 mL/min but noticeable effect on a 50x2 mm column with 0.2 mL/min.
Will RT increase or decrease if length of heat exchange capillary is increased?
Increased lenght means increased extra column volume, therefore it will also increase the RT.
I suppose what SamNas might also be hinting at is whether the actual temperature-pre-equilibration will be affected by the length of the heat-exchange capillary, and what that will do to retention times.

I have no idea, but at 2mL/min the time spent by solvent in a 3uL capillary is less than a tenth of a second, so I suppose it's possible the solvent isn't going to be fully equilibrated?
5 posts Page 1 of 1

Who is online

In total there are 22 users online :: 2 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry