Problem with active compound elution
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
15 posts
Page 1 of 1
I had good peaks before and then doubling of peaks occurred in my chromatograms. I washed the column and now only solvent peak elutes while no active compound elutes. I am using Agilent 1200 series hplc. Using same method only length of heat exchange capillary is changed.
-
- Posts: 273
- Joined: Mon Aug 04, 2014 8:13 am
Hello
More details please...
-compound details
-solvents type
-eluents
-column
-method parameters
-chromatogram
Regards
Tomasz Kubowicz
More details please...
-compound details
-solvents type
-eluents
-column
-method parameters
-chromatogram
Regards
Tomasz Kubowicz
-
- Posts: 13
- Joined: Wed Sep 17, 2014 7:01 am
Compound is Dexamethasone and Prednisolone
solvent methanol,
mobile phase H2O and ACN ( Gradient)
Column C-18
Temp 30C , Time 8 mins,
I m trying to post chromatogram but couldnt get it done
solvent methanol,
mobile phase H2O and ACN ( Gradient)
Column C-18
Temp 30C , Time 8 mins,
I m trying to post chromatogram but couldnt get it done
-
- Posts: 273
- Joined: Mon Aug 04, 2014 8:13 am
Hello
Possible problems:
-dead volume (mixing chamber) can cause double peak. Check all connections
-sample solvent (100%methanol) is stronger than mobile phase. It can cause double peak
-you mentioned that you have no peak after column cleaning. Wash your column with methanol and with your initial mobile phase.
Regards
Tomasz Kubowicz
Possible problems:
-dead volume (mixing chamber) can cause double peak. Check all connections
-sample solvent (100%methanol) is stronger than mobile phase. It can cause double peak
-you mentioned that you have no peak after column cleaning. Wash your column with methanol and with your initial mobile phase.
Regards
Tomasz Kubowicz
-
- Posts: 927
- Joined: Mon Feb 13, 2006 6:28 pm
What do you mean by "doubling of peaks" ?I had good peaks before and then doubling of peaks occurred in my chromatograms. I washed the column and now only solvent peak elutes while no active compound elutes. I am using Agilent 1200 series hplc. Using same method only length of heat exchange capillary is changed.
If you meant what I mean then you have voids in the column and you may say bye-bye to it.
-
- Posts: 24
- Joined: Thu Aug 28, 2014 11:55 pm
Back flush your column with ACN, sounds like you have a blocked frit
-
- Posts: 13
- Joined: Wed Sep 17, 2014 7:01 am
Thank you all. Keeping in view your suggestion I have back flushed the column. But still only a small peak at 0.5 sec elutes ( I suppose it is solvent peak). No other signal appears and the baseline is almost flat. How can I know if the column needs to be replaced.
-
- Posts: 421
- Joined: Wed Jul 18, 2012 6:45 am
Along with all columns comes a certificate. There you will find a sample chromatogram with a standard mix. Use the compounds of this mix and run them under the same conditions, then you should get the same chromatogram. Of course not the SAME chromatogram, but you will see if your plate count is close to the one stated, and your retention times are similar.
It is a good idea to create an own test mixture you run from time to time to see if your column performance varies. Of course this only helps if this test mix is run immediately after purchasing the column.
It is a good idea to create an own test mixture you run from time to time to see if your column performance varies. Of course this only helps if this test mix is run immediately after purchasing the column.
-
- Posts: 13
- Joined: Wed Sep 17, 2014 7:01 am
Thankyou
-
- Posts: 13
- Joined: Wed Sep 17, 2014 7:01 am
The peak for solvent is smaller but the retention time is same. Only that no other compound is being detected.
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
At 0.5 sec it most certainly is *not* a peak at all but rather the injection mark. You did not specify the column volume or the flow rate, but you can estimate the t0 (the time for the "solvent" peak) atBut still only a small peak at 0.5 sec elutes ( I suppose it is solvent peak).
0.5 * L * dc^2 / F , where L is the column length, dc is the column internal diameter, and F is the flow rate.
If this were my problem I would do the following until I found the source of the difficulty:
1. Confirm that the flow rate is accurate (run the outlet into a 10 mL graduate cylinder and time how long it takes to fill)
2. Confirm that the detector is functioning (an Agilent 1200 should have a self-test function) and set at the correct wavelength
3. Confirm that the autosampler is functioning (observe the tray and syringe; listen for stepper motor operation; measure volume in a vial before and after injecting a reasonably large volume.
4. Confirm that the gradient system is functioning properly by running the gradient step-test (http://www.lcresources.com/resources/TSWiz/hs450.htm)
5. Prepare a fresh batch of mobile phase
6. Test with a new column
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 13
- Joined: Wed Sep 17, 2014 7:01 am
I ll surely check for all the steps today
-
- Posts: 13
- Joined: Wed Sep 17, 2014 7:01 am
The flow rate is faulty
.When i set it to 1ml/min, the volume collected is only about 0.6 ml/min. Pump A is giving accurate flow but pump B is not

-
- Posts: 512
- Joined: Tue Jul 19, 2011 9:26 pm
Pure ACN on pump B?
-
- Posts: 13
- Joined: Wed Sep 17, 2014 7:01 am
Thank you very much for your guidance. I sonicated the filter first with H2O and then with methanol and viola!!! problem is solved.
15 posts
Page 1 of 1
Who is online
In total there are 45 users online :: 2 registered, 0 hidden and 43 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot], Google [Bot] and 43 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot], Google [Bot] and 43 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science