When do we need vent gcms
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
3 posts
Page 1 of 1
gold seal change? or column change? if i do not need vent the system for gold seal, do I need to cap the column by using septum seal?
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
For changing the gold seal, just cool oven to ambient and to the work, pulling that little air through the column won't damage it if you will let it flush for 10 minutes before heating the oven up again. Don't plug it with a septa, as it can suck fragments down into the column because of the vacuum.
If you are going to change the column you normally vent the MS since such a large air leak from the removed column can cause strain on the high vacuum system. I have before changed columns by venting cooling the source and shutting down the turbo pump but leaving the rough pump running and doing a quick one minute switching of columns, but it isn't easy and if you haven't done it before probably not the best idea to try.
If you are going to change the column you normally vent the MS since such a large air leak from the removed column can cause strain on the high vacuum system. I have before changed columns by venting cooling the source and shutting down the turbo pump but leaving the rough pump running and doing a quick one minute switching of columns, but it isn't easy and if you haven't done it before probably not the best idea to try.
The past is there to guide us into the future, not to dwell in.
-
- Posts: 206
- Joined: Sun Oct 09, 2011 5:08 pm
Thanks so much!
For changing the gold seal, just cool oven to ambient and to the work, pulling that little air through the column won't damage it if you will let it flush for 10 minutes before heating the oven up again. Don't plug it with a septa, as it can suck fragments down into the column because of the vacuum.
If you are going to change the column you normally vent the MS since such a large air leak from the removed column can cause strain on the high vacuum system. I have before changed columns by venting cooling the source and shutting down the turbo pump but leaving the rough pump running and doing a quick one minute switching of columns, but it isn't easy and if you haven't done it before probably not the best idea to try.
3 posts
Page 1 of 1
Who is online
In total there are 7 users online :: 1 registered, 0 hidden and 6 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Baidu [Spider] and 6 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Baidu [Spider] and 6 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science