I know DMSO is not an ideal diluent for HPLC. It will distort the peak shape. Is there any theory or explaination about it? Thank you in advance.
-
- Posts: 216
- Joined: Mon Apr 04, 2005 12:26 am
- Location: TN
Don't know for sure, but probably just too strong a solvent and causes peak to move with DMSO before actually interacting with column/eluent.
We find DMF is a very good replacement for DMSO. Usually better to make up more concentrated solution and inject 5 ul than to prepare a diluter solution and inject 25 uls.
THF not a bad solvent, but not as good as DMF or DMSO for many of our compounds.
Also use TFA as a solvent. However, can't let compounds with alcohols sit for long time because converts some alcohols to TFA esters in about 1 hour. So prepare and inject fresh. Also, leave the tops off the bottles, because after one injection will dissolve silicones in top and give a lot of mass spec background ions.
We find DMF is a very good replacement for DMSO. Usually better to make up more concentrated solution and inject 5 ul than to prepare a diluter solution and inject 25 uls.
THF not a bad solvent, but not as good as DMF or DMSO for many of our compounds.
Also use TFA as a solvent. However, can't let compounds with alcohols sit for long time because converts some alcohols to TFA esters in about 1 hour. So prepare and inject fresh. Also, leave the tops off the bottles, because after one injection will dissolve silicones in top and give a lot of mass spec background ions.
Sailor
-
- Posts: 480
- Joined: Fri Sep 03, 2004 7:57 am
- Location: Indonesia
Are both safe enough for seal and PEEK?
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
- Location: Alamo, CA / USA
PEEK is incompatible with THF.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 835
- Joined: Thu Apr 14, 2005 7:00 am
Jan,
what type of peak distortion do you get?
what type of peak distortion do you get?
-
- Posts: 17
- Joined: Mon Aug 30, 2004 7:55 pm
Actually I never used more than 10% of DMSO in my sample so I did not have a peak shape issue. What I do is that I dissolve the sample with samll amount of DMSO then dilute it with my mobile phase. I just heard from other people that if you use only DMSO as a diluent and you will see bad peak shape. Anybody has real experience about that?
-
- Posts: 144
- Joined: Mon Aug 30, 2004 10:12 pm
If the injection volume is small, 5 or 10 uL, DMSO is no problem. It is our preferred saponification solvent, so we have done bazillions of samples with DMSO.
It is a very "strong" solvent and viscous. I expect both these factors are a problem with large volume injections.
It is a very "strong" solvent and viscous. I expect both these factors are a problem with large volume injections.
Bill Tindall
7 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science