Agilent 6890 detector output not displaying

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
We have a 6890 dual GC. Both detectors appear to be working fine on the instrument itself (have a non-zero output). However, on the computer program, the signal 2 says N/A on the window where the spectrum is recorded. Signal 1 says "front detector" and I can see the detector output in real time. Any idea on how to get signal 2 (the back detector) to show up on the computer?

Also, if I try to run a sample on just the front detector, the FID file isn't being recorded. I'm guessing these problems might be related. Thanks!
Designate back detector as Signal 1.
...However, on the computer program, the signal 2 says N/A on the window where the spectrum is recorded. Signal 1 says "front detector" and I can see the detector output in real time. Any idea on how to get signal 2 (the back detector) to show up on the computer?
Did you resolve your problem? What software are you using to analyze the data? (ex: Chemstation, etc,)

Also, if I try to run a sample on just the front detector, the FID file isn't being recorded. I'm guessing these problems might be related. Thanks!
You need to go through the analytical method's parameters, ensuring that the injection source, signals, and data integration parameters are all set properly. Do you actually use both front and rear columns together for dual analyses?
Assuming you run Chemstation
However, on the computer program, the signal 2 says N/A on the window where the spectrum is recorded. Signal 1 says "front detector" and I can see the detector output in real time. Any idea on how to get signal 2 (the back detector) to show up on the computer?
I

In the method settings, under signal, select signal 2 as FID back

Also, if I try to run a sample on just the front detector, the FID file isn't being recorded. I'm guessing these problems might be related. Thanks!
Under method settings, under signal, check window to store signal
4 posts Page 1 of 1

Who is online

In total there are 25 users online :: 0 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 25 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry