We want to determine active ingredient dissolved in dissolution medium (HCl 0.1N). The system as follows:
Column: 150 x 4.6 mm, contains C18 column packing.
Mobile phase: Phosphate buffer, pH 8.0 – acetonitrile (73:27)
Flow rate: 1.5 mL/minute
Column temperature: 55 C
Detection: UV 210 nm
Injection volume: 50 uL (1 – 15 ug/mL)
After several hours, the retention time still unstable (tend to decrease). Here is the retention of last 25 injections:
6.959
6.955
6.918
6.886
8.899
6.877
6.844
6.857
6.837
6.834
6.833
6.800
6.769
6.755
6.758
6.758
6.725
6.753
6.715
6.688
6.673
6.660
6.658
6.618
6.575
We use the same system in assay for tablets, but in lower injection volume (10 uL) and higher working concentration (100 ug/mL). Solvent in assay is mobile phase. The problem is not occurred in assay.
Any explanation please why the retention could not be stable…