I'm working on USP Method verification's. My Question is regarding RI Detectors and their Temperature maintenance. there is no RI Detector Temperature criteria for Sucralose, NF Assay method (USP37-NF32). when we were doing the method verification for this method, we kept the detector temperature ambient. But, the detector was not at all stabilizing. so we opted to 35°C and we got good results with repeat-ability and reproducibility. But in <621> system Suitability there is no criteria for RI Detector Temperature variation. So we contacted USP and their answer was to validate this method. Do i need to validate this method or verification is good?