Can anybody tell me the impact of Carbon loading in the column on the separation of the peak?

What happen when the Carbon loading is High?

What happen when carbon loading is Low?

Is the high carbon loaded column is suitable for Impurity separation?



Aruna

All other things being equal, a higher carbon load will result in longer retention times.

The problem is that all other things are seldom equal. Retention and selectivity are controlled by things like surface area, pore size distribution, surface coverage, and bonded-phase chain length, which means that it is quite possible to have two packings with similar carbon load (% carbon), but vastly different chromatographic characteristics.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 17 users online :: 1 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: John Guajardo and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry