- Posts: 4
- Joined: Thu Jul 24, 2014 4:07 pm
I've had problems for a while with auto-tuning, and the instrument has been unable to perform detector gain calibration. I have repeatedly cleaned the ion source, lens assembly, prefilter etc, and also replaced the electron multiplier (temporarily fixed the problem for ~1 month), topped up calibration gas vial, and replaced the filament. Apart from this issue the GC-MS has been working fine. No faults were previously detected with the diagnostic tests, but yesterday it failed on lens diagnostics. Lens 3 appears to be the issue. It does not show a flat line response, but instead starts off flat and then ramps linearly before leveling off again. I cleaned the lenses, but this had no effect - apart from the auto-tune (inc. detector gain calibration) now seems to work OK for the first time in 3 months!! The troubleshooting section of the DSQ manual suggests that the analog PCB or regulator PCB in the power module may need replacement. Question is how do I know which (if either!) is causing the lens problem? I'd be really grateful for any suggestions.....