Mr.James,
It is new agilent GC 7890B & ms 5977 six month back only we purchased. previously we didn't face this much high response decrease within one day last 2 week on words we are facing this problem (after source clean). but after source clean we got very good response for PFTBA 219 ion nearly 140%. now today i did Etune again the now we are getting PFTBA 213 ion response only 80%. Slight variation in EM voltage also but very less.
our analysis is SIM mode we are monitering ions from 200 to 250 m/z only.
Regards
Ashok
With the drop in 219 percentage it could be an air leak causing oxidization of the source over time, or there could be a problem with the column having high bleed and making the source dirty quickly, it has been known to happen that you get a bad column though it is rare.
If you set it to scan from 18m/z up to 510m/z you will see if there is any air and water getting into it, sometimes just scanning like that shows more than the air water check tune does.
Also try scanning with the cal gas valve closed and threshold set to about 10 and see how much background you have. You should only have traces of 18, 28, 32, 40, 44 and maybe a small amount of 207 from the column bleed, but not much else.
If you do see air and water try tightening the column interface fitting about 1/4 turn, as those ferrules can shrink after they are first installed.