Area is continuously decreasing in GC-MS
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
7 posts
Page 1 of 1
Hi all this is ashok from india. I hope all are good. Here in our GC-MS we are running standards every day at least 10 injection.The area of standard decreasing day to day let see today I am getting 30000 area for 50 ppb STD but the next day i am getting 29000 and the next day (third day) i am getting 28000 area. Let me know what may be the reason? Please help me.....
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
Air leaks can degrade sensitivity over time and also if you inject aqueous samples you can lose sensitivity. Column temperature too high can cause excessive column phase bleed which will also cause lower sensitivity over time.Hi all this is ashok from india. I hope all are good. Here in our GC-MS we are running standards every day at least 10 injection.The area of standard decreasing day to day let see today I am getting 30000 area for 50 ppb STD but the next day i am getting 29000 and the next day (third day) i am getting 28000 area. Let me know what may be the reason? Please help me.....
Usually lower sensitivity comes from changes in vacuum or a dirty source if it is that fast. Longer periods of time it can just be normal wear of the electron multiplier.
The past is there to guide us into the future, not to dwell in.
-
- Posts: 19
- Joined: Sun May 18, 2014 12:28 pm
Mr.James
Thanks for your comment. But we checked instrument leak also there is no leak in instrument and we are using organic samples(Hydrocarbons).there is no column bleed ions in the TIC and we are using same method in other instrument there we are getting very good repeatable result so do you think problem in oven temperature?.
Recently service engineer cleaned the source so we are suspecting it is not the problem of ion source. we need to monitor our vacuum system from today on words..
So shall i wait for some more days? then it will automatically become normal?
Thanks
Ashok
Thanks for your comment. But we checked instrument leak also there is no leak in instrument and we are using organic samples(Hydrocarbons).there is no column bleed ions in the TIC and we are using same method in other instrument there we are getting very good repeatable result so do you think problem in oven temperature?.
Recently service engineer cleaned the source so we are suspecting it is not the problem of ion source. we need to monitor our vacuum system from today on words..
So shall i wait for some more days? then it will automatically become normal?
Thanks
Ashok
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
How old is the electron multiplier? If it is near the end of its life, it can give these symptoms, slowly losing sensitivity over short periods of time.
What model and manufacturer of MS do you have?
What model and manufacturer of MS do you have?
The past is there to guide us into the future, not to dwell in.
-
- Posts: 19
- Joined: Sun May 18, 2014 12:28 pm
Mr.James,
It is new agilent GC 7890B & ms 5977 six month back only we purchased. previously we didn't face this much high response decrease within one day last 2 week on words we are facing this problem (after source clean). but after source clean we got very good response for PFTBA 219 ion nearly 140%. now today i did Etune again the now we are getting PFTBA 213 ion response only 80%. Slight variation in EM voltage also but very less.
our analysis is SIM mode we are monitering ions from 200 to 250 m/z only.
Regards
Ashok
It is new agilent GC 7890B & ms 5977 six month back only we purchased. previously we didn't face this much high response decrease within one day last 2 week on words we are facing this problem (after source clean). but after source clean we got very good response for PFTBA 219 ion nearly 140%. now today i did Etune again the now we are getting PFTBA 213 ion response only 80%. Slight variation in EM voltage also but very less.
our analysis is SIM mode we are monitering ions from 200 to 250 m/z only.
Regards
Ashok
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
With the drop in 219 percentage it could be an air leak causing oxidization of the source over time, or there could be a problem with the column having high bleed and making the source dirty quickly, it has been known to happen that you get a bad column though it is rare.Mr.James,
It is new agilent GC 7890B & ms 5977 six month back only we purchased. previously we didn't face this much high response decrease within one day last 2 week on words we are facing this problem (after source clean). but after source clean we got very good response for PFTBA 219 ion nearly 140%. now today i did Etune again the now we are getting PFTBA 213 ion response only 80%. Slight variation in EM voltage also but very less.
our analysis is SIM mode we are monitering ions from 200 to 250 m/z only.
Regards
Ashok
If you set it to scan from 18m/z up to 510m/z you will see if there is any air and water getting into it, sometimes just scanning like that shows more than the air water check tune does.
Also try scanning with the cal gas valve closed and threshold set to about 10 and see how much background you have. You should only have traces of 18, 28, 32, 40, 44 and maybe a small amount of 207 from the column bleed, but not much else.
If you do see air and water try tightening the column interface fitting about 1/4 turn, as those ferrules can shrink after they are first installed.
The past is there to guide us into the future, not to dwell in.
-
- Posts: 2
- Joined: Tue Nov 18, 2014 10:13 am
Hi Ashok,
did you solve the problem and could tell us what parameters were responsible causing the problems?
I have quite the same problems...
Dennis
did you solve the problem and could tell us what parameters were responsible causing the problems?
I have quite the same problems...
Dennis
7 posts
Page 1 of 1
Who is online
In total there are 28 users online :: 3 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot], Google [Bot], Semrush [Bot] and 25 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot], Google [Bot], Semrush [Bot] and 25 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science