Chemstation aborts after injection
Posted: Mon Jul 07, 2014 10:16 pm
by sabouelm
We have a 1100 Agilent HPLC. It's been working fine, but recently it started aborting the run directly after injection. The logbook says that it is due to an error (but no error information/number). We tried remaking the method and even reinstalled chemstation, but the error persisted! the run is aborted after injection.
Do you have any suggestions/tips to solve such problem?
Re: Chemstation aborts after injection
Posted: Wed Jul 09, 2014 5:20 pm
by Consumer Products Guy
Set the vial number to nothing for a blank injection, then run the Method, see if same happens.
During a real injection, I'd also look at that moment and see if the pressure drops out.
Re: Chemstation aborts after injection
Posted: Fri Jul 11, 2014 3:56 pm
by djacqo
Have you loaded the logbook from the data folder from the sequence (rather than the current logbook)? I've noticed that sometimes has more information on which module shutdown.
Do you have any modules that are connected by CAN but not in the current configuration?