Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
We have a 1100 Agilent HPLC. It's been working fine, but recently it started aborting the run directly after injection. The logbook says that it is due to an error (but no error information/number). We tried remaking the method and even reinstalled chemstation, but the error persisted! the run is aborted after injection.

Do you have any suggestions/tips to solve such problem?
Set the vial number to nothing for a blank injection, then run the Method, see if same happens.

During a real injection, I'd also look at that moment and see if the pressure drops out.
Have you loaded the logbook from the data folder from the sequence (rather than the current logbook)? I've noticed that sometimes has more information on which module shutdown.
Do you have any modules that are connected by CAN but not in the current configuration?
3 posts Page 1 of 1

Who is online

In total there are 33 users online :: 0 registered, 0 hidden and 33 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 33 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry