Hello. I am new to this forum, so please forgive me if I have posted this question in the wrong place.

I am trying to create a method on GC/FID for detecting n-dodecane in dichloromethane. My problem is that I am getting a huge solvent peak and not a very accurate analyte peak. Right now, I have a 0.1 n-dodecane in DCM sample, He carrier gas, injecting 0.5ul sample, 10psi flow rate, temp starting at 40C, ramped at 10C/min to a max of 225C.

Is there a way to adjust the method so the solvent peak is smaller, or just start reporting after the solvent comes off the column so that I don't have a giant peak at the beginning of my run that dwarfs everything else?

Thank you in advance for any advice you can give me (even if it's post this thread in a different place)

-Kyle
Increase split ratio , use a larger diameter inlet liner , and more head-pressure as a starter.
Does your chromatography data system allow a solvent delay? It should allow you to start data collection at a specified time after the run start.
kylebenton,

Splitless or split or modified packed? What kind of column (especially dimensions - usually a plate on the column itself)? 0.1 what for concentration? What kind of data system are you using?

Best regards,

AICMM
4 posts Page 1 of 1

Who is online

In total there are 14 users online :: 1 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry