How soon after pump down is completed can you heat up the MSD on agilent 5973N?
-
- Posts: 927
- Joined: Mon Feb 13, 2006 6:28 pm
ImmediatelyHow soon after pump down is completed can you heat up the MSD on agilent 5973N?
-
- Posts: 3
- Joined: Sat May 31, 2014 10:03 pm
thanks.
I am experiencing high air/water leak, ~6.03% & 18.87% respectively. I isolated the MS and re-check the a/w. It was about 7 & 10%. With the column re-attached the a/w was about 6, 12%. I baked out the MS via bake out program in chemstation, the a/w was about the same, 6 & 15%. However, the instrument I baked out the gc at 260 degrees C for ~1 hr. Roughly the same results for a/w. I checked for leaks at the door to the ms, between the ms and gc, at the x-fer line nut, the connection between the x-fer line and analytical column, the connection between the guard column and analytical column, the column nut and the retainer nut at the end of the end of the inlet body (which keeps loosening up). No leaks indicated.
If I have run the instrument with the leak, I get a few shots before I have to open up the ms to change the filament.
I refilled the calibrant vial recently, however the leak was present before but not as bad. Could that be a factor? Are there other problems that give a high a/w check, that may not truly be a leak? Is there another place in the system I should be looking for a leak?
I am experiencing high air/water leak, ~6.03% & 18.87% respectively. I isolated the MS and re-check the a/w. It was about 7 & 10%. With the column re-attached the a/w was about 6, 12%. I baked out the MS via bake out program in chemstation, the a/w was about the same, 6 & 15%. However, the instrument I baked out the gc at 260 degrees C for ~1 hr. Roughly the same results for a/w. I checked for leaks at the door to the ms, between the ms and gc, at the x-fer line nut, the connection between the x-fer line and analytical column, the connection between the guard column and analytical column, the column nut and the retainer nut at the end of the end of the inlet body (which keeps loosening up). No leaks indicated.
If I have run the instrument with the leak, I get a few shots before I have to open up the ms to change the filament.
I refilled the calibrant vial recently, however the leak was present before but not as bad. Could that be a factor? Are there other problems that give a high a/w check, that may not truly be a leak? Is there another place in the system I should be looking for a leak?
-
- Posts: 880
- Joined: Fri Jan 04, 2008 3:03 pm
I had a similar issue not that long ago and it was the O-ring thet seals the transfer line assembly to the MS box. It's kind of a pain to get to.
-
- Posts: 3
- Joined: Sat May 31, 2014 10:03 pm
Thanks bigbear. I will look into that.
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
Also check the screws on the end plates, I have had to tighten those before. The four big Torx head ones at the front and back.
The past is there to guide us into the future, not to dwell in.
6 posts
Page 1 of 1
Who is online
In total there are 38 users online :: 3 registered, 0 hidden and 35 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Amazon [Bot], Google [Bot], per_oxid and 35 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Amazon [Bot], Google [Bot], per_oxid and 35 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science