ThermoFIsher - Power outage and GC MS ITQ 700
Posted: Thu May 15, 2014 10:40 am
by klam
Good afternoon,
Here is the issue : We do have sometimes power outage and I'm really afraid to run the Themo GS MS FocusMS - ITQ700.
Is there a way for the MS to shut down properly in case of poweroutage? We are going to add a UPS in the future .
Re: ThermoFIsher - Power outage and GC MS ITQ 700
Posted: Thu May 15, 2014 3:32 pm
by christophe_j
what about a backup power battery? this was should the power drop you can shut down while battery keeps keeps running the MS
Posted: Thu May 15, 2014 3:34 pm
by klam
Yes, we are thinking about that but this is only ok for a short period of time and this will prevent the GCMS to run during the night...
We will have a power backup but we need to install a triphasic power outlet first...
Posted: Thu May 15, 2014 7:21 pm
by leadazide
Usually the power outage itself is not the issue.. It is the uncontrolled return of power that usually causes problems..
I would install a switch that automatically disengages the power is there is a power outage.. The when you know the power is back you can manually engage the power in a controlled manner..