Does this happen for every level of the calibration for the drugs that are affected ? If not, which levels are affected ? How does the linearity (r-squared) look ?

Peter
Peter Apps
I second Peter Apps' questions. It sounds to me like the calibration is not quite linear. But I'd want to know more details before really making a guess as to what's going on.

It depends on what regulations you're under, but some different things you try do are limit the calibration range (if your calibration is 1ppm - 100ppm, but all samples are below 20ppm, get rid of all cal points above 20), change the weighting of the calibration curve to 1/x or 1/x^2, or simply use a curve that fits better than linear. As I'm in a research environment, I just go with the latter and pick the curve that gives the best fit. In my view, that's the most accurate, and better than messing with data until you can slap a linear curve on everything.

As an aside, sometimes a good R^2 value can be misleading. You can have a curve that appears to be linear at first glance and gives a good R^2, but is very off at low levels. Very relevant if you are doing any trace analysis where your samples tend to be at the low end of the cal.
THANKS FOR for the replies, I was able to get help from outside the forum to solve issues on the problem compounds.
Hello,

I suppose you're using the esi probe... could you tell me what source setting did you choose for drug of abuse analysis?

thanks
when I run my curves, the calculated conc for the calibrators is significantly less than the the theoretical conc.
This sounds very much like either you are not using the right model for your calibration curve, or that something (another analyte, instability, binding to walls of autosampler vial, etc.) interferes with the analysis of the problematic drug.

Try running a calibration curve with only the troublesome drug. Check residuals to make sure it is linear throughout. R^2 is nearly worthless for this if your calibration range is >2 orders of magnitude. If it isn't linear, you need to figure out why.
All standard disclaimers apply. My posts are my opinions only and do not necessarily reflect the policies of my employer.
6 posts Page 1 of 1

Who is online

In total there are 29 users online :: 0 registered, 0 hidden and 29 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 29 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry