I am currently having a puzzling problem with ZIC-HILIC and ZIC-cHILIC columns.Merck has two application notes for this:
on the HILIC column they suggest running isocratic 70:30 ACN/ammonium acetate pH 4 100 mM
on the cHILIC they suggest isocratic 70:30 ACN/ammonium acetate pH 4.5 100 mM
We purchased one of each (100 A pore x 4.6 x 150 mm x 3.5 or 3 µm respectively), and one appropriate guard column for each (2.1 x 50 mm x 3.5 or 3 µm). For the first use, we primed our system with the solvents of the application note, connected the guard, flushed for 20 mins at 1 ml/min, connected the analytical column through the column coupler that comes with the guard, and ran a series of standards diluted in mobile phase at a flow of 2 ml/min. Mobile phase was prepared with freshly opened HPLC grade water and ACN from Biosolve and HPLC grade salt and acetic acid from Sigma. The pH was checked over an aliquot taken out of the bottle to prevent introducing contamination from the probe. Solvent were visually checked for dust or particulate and no further filtering was performed.
In both cases the first injection of AMP, ADP and ATP gave excellent peak shapes and the k' described in the application, but immediately following the weird problems started...
In the case of the HILIC column, peak shape started to deteriorate rapidly. The peak tailing became so bad that after 5 injection ATP was no longer detectable. Therefore we decided to rinse the column and the guard column separately, following the procedure described in the booklet. Rinse was 20 volumes of 1ml/min water, followed by 20 volumes NaCl 0.5 M, and finally 40 volumes water. This did not help, as afterwards the peak shape was even worse. We run the samples with the guard column, and without. The result was the same. We tried to reverse the flow (0.5 ml/min, 40 volumes) and then analyze the standards again in normal direction: the peak was split in two.
In the case of the cHILIC, not only peak shape degraded, but we had weird backpressure problems. Peak tailing started to show up at the second run, and between the second and the third run the pressure got over 350 bar and the system stopped. After stopping the pump, the pressure did not drop. Therefore we proceeded to disconnect the fitting at the inlet side of the analytical colum. The pressure relieved. I thought of a misalignement problem of the fitting, and changed the column coupler with a homemade one. Peak shape was still tailing, and after two runs the backpressure problem showed up again. We changed guard column and coupler, and peak tailing was still there. We ran the column without guard, and tailing persisted. Lookign accuratley at the chromatogram, it looks more like a broad shoulder after every peak rather than tailing.
If I look at the data I have, it leads to the conclusion of a clogged frit or dislodged bead packing. This leaves me perplexed however, as I cannot recognize any step in which I did something wrong. I normally run HPLC-TOF on RP columns and I know how to prevent solvent contamination. On the same system used for the ZIC-HILIC I regularly use 2.6 µm columns and never had a problem with clogging.
Has anyone else faced the same problem?