Now I have a new (well, used, but new to me) ELSD detector: Varian LC-380 from 2008. First thing I tried is: water, 1 ml/min at 40° nebulizer, 80° evaporator and 1.4 slm gas flow. Noise is 2.6 mV at gain setting 1. This is far away from the specification 0.5 mV. Even without flow I have a noise level of about 0.7 and drift of -0.6 mV over 5 min.
Next things to try are another batch of water and pure acetonitrile (specifications are <0.3 mV noise).
About sensitivity: I tried to inject 10 µl 10 mg/L caffeine (equivalent to 100 ng). This should be easily detectable. I only see flat line. 50 mg/L is a peak I would say OK this is about my LOQ. Parameters used are acetonitrile/water 25/75 at 1 ml/min, 40°C /40°C /1.4 slm.
I'd appreciate any help in getting more familiar with this device.
Thanks a lot
Jörg
PS: I tried the cleaning procedure from the manual: 4 h acetone/water 50/50 at 2 ml/min and 2.8 slm flow. No change in noise.