Hi,
In order to validate the linearity of an anlytical method for determination of tyrosine in human plasma, i used spike-and-recovery experiment .
My question is : after injection in hplc of standards solutions ( range of 1 to 20 mg/l) . i check the linearity of the curve , R2=0,999 and the equation of regression line : y=1.9 x+ 0.3.

Should i calculate the concentration of spiked plasma using the equation of standard solutions .

Thank You.
Should i calculate the concentration of spiked plasma using the equation of standard solutions .
How else would you do it? :wink:

If this were my problem, I would be somewhat concerned about what looks to be a significant y-intercept. What was the standard error?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Pandit.11 and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry