Problem with column active heater of HClass System

Problems instrument:
- the column manager does not respond to temperature regulation bringing the column to maximum temperature.

Software problems:
- Empower records only the room temperature;
- Empower does not allow to change the temperature of the column because the Column Manager section is not reported;
- Empower acquires only the data of "Sample Manager" and of "Quaternary Solvent Manager".

Any help would be appreciated. Thanks
Hi,

Sounds like you mightn't have the correct options selected for the column manager. Should have the same or compatible firmware on all modules.
When you open the acquity console you should see all modules, then you can verify firmware. Also, you should see the cm and be able to set the temp, on the two channels.
Do you have a service contract for the instrument?

Regards,
Bull76
2 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry