How do you go about setting the flows for a micropacked column in a 6890 GC? I'm currently still trying to analyze the amount of CO in a CO2 stream using a 0.53 mm ShinCarbon ST micropacked column with a split/splitless inlet and a TCD with helium as carrier gas, but it's not working well.
When I installed the column, I calibrated the column flow vs. the inlet pressure using a flow meter and I got this calibration curve.

The recommended flow rate for this micropacked column is 3 to 5 mL/min, so I chose 5 mL/min (for no particular reason at all). That means I have to set my inlet at 4.06 bar.
Okay, now the question becomes, how do I set the split ratio?
I thought then that given that
Total flow = Septum purge vent + column flow + split vent
and that the purge vent and column flows are fixed automatically by the EPC and by the inlet pressure, respectively, the split ratio is controlled only by the total flow to the inlet. Is this correct?
So, if I set a total flow of 33 mL/min at 4.06 bar, 5 mL/min will go to the column and 3 mL/min will go to the septum purge, leaving a split vent flow of 25 mL/min, that is, a 1:5 split ratio.
Does this make sense?