Sorry I forgot to write: widths are calculated at the baseline.If the widths you use are at half height then 50 % recovery (in half the peak volume) is pretty good accuracy.
Peter
I am using Dionex HPLC system with Chromeleon software. Column is HILIC with 3 µm particles (150x4.6 mm) and detector I´m using is fluorescent detector.You're gonna first have to tell us what system, software you're using, how you determined recovery, etc
I injected 100 ul. And I collected fraction that has as you said volume of about 300 ul. Then I dried the sample down to have again the same volume as the injected sample-to 100 ul. And then I injected whole volume.Your injection volume is one of the many details that you do not provide, but let's assume that it is 150 ul. At the flow rates you give, presuming that you collect for the whole of the peak width, your fraction volume is 300 ul - so your substance comes out of the column in double the volume it went in, and your collected fraction has a concentration one half of that of the original solution. Then you inject 150 ul of the fraction, which contains half as much substance as the first injection, and get a peak that has half the area of the original peak. You then calculate that you have 50% recovery.
Peter
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.